Calibrating a pressure switch involves adjusting its set point(s) to ensure it activates (or deactivates) at the desired pressure level. Here's a general guideline, but remember to consult the specific user manual for your pressure switch model for precise instructions and safety precautions:
Safety Precautions:
-
- Pressure Relief: Before starting, ensure the pressure source is isolated and depressurized. Release any built-up pressure in the system following the recommended safety procedures.
-
- Eye Protection: Wear safety glasses to protect your eyes from potential debris or accidental splashes during the process.
Materials:
-
- Pressure gauge with a higher accuracy than the pressure switch
-
- Hand pump or compressed air source (ensure compatibility with the pressure switch)
-
- Multimeter (for switches with electrical outputs)
-
- Adjustable wrench or screwdriver (for set point adjustment)
Steps:
- Disconnect and Isolate: Disconnect the pressure switch from the process piping and isolate it from the pressure source.
- Connect the Pressure Source: Connect the pressure source (hand pump or compressed air) to the pressure switch inlet port.
- Connect the Pressure Gauge: Connect the pressure gauge to the switch's output port or a separate pressure tap on the line for a more accurate reading.
- Electrical Connection (if applicable): For switches with electrical outputs, connect the multimeter to the switch terminals according to the manufacturer's instructions (usually measuring continuity).
- Observe Initial State: Observe the initial state of the switch (open or closed) and the corresponding pressure reading on the gauge (usually indicated in the manual).
- Increase Pressure: Slowly increase the pressure using the hand pump or compressed air source while monitoring the pressure gauge.
- Set Point Adjustment: For normally open (NO) switches, watch for the switch to change state (become electrically closed or show continuity on the multimeter) as the pressure reaches the desired set point. Use the adjustment mechanism (refer to the manual) to fine-tune the set point until the switch activates at the correct pressure according to the gauge reading.
- Repeat for Normally Closed (NC) Switches: If your switch is normally closed (NC), follow steps 6 and 7 but observe when it switches to open (continuity breaks on the multimeter) and adjust for the desired deactivation pressure.
- Decrease Pressure (Optional): Some pressure switches allow for setting a separate deadband (the pressure difference between activation and deactivation). You can repeat steps 6-8 while decreasing pressure to set the deactivation point.
- Record Settings: Record the final set point pressure readings for future reference.
- Reconnection: Once calibration is complete, carefully reconnect the pressure switch to the piping and ensure all connections are secure.
- System Pressure Re-establishment: Slowly re-introduce pressure into the system while monitoring for proper operation and leaks.
Additional Tips:
-
- Consult the user manual for specific instructions on the adjustment mechanism and any limitations of your pressure switch model.
-
- If the pressure switch has tamper-proof seals, avoid breaking them to maintain warranty and safety standards. In such cases, consult a qualified technician for calibration.
-
- It's recommended to perform pressure switch calibration periodically, especially in critical applications, to ensure continued accuracy.
Remember, if you're unsure about any steps or safety procedures, it's always best to consult a qualified technician for assistance.