Calibration of Pressure Switches

1 Posts
1 Users
0 Reactions
230 Views
Posts: 18330
Admin
Topic starter
(@click2electro)
Member
Joined: 4 years ago

Calibration of pressure switches is a critical process to ensure their accuracy and reliability in detecting and controlling pressure levels within a specified range. Here's a general overview of the calibration process for pressure switches:

  1. Gather Equipment: Prepare the necessary equipment for calibration, including a pressure calibrator or deadweight tester, a pressure source capable of generating the desired pressure range, and any accessories required for connection.

  2. Verification of Calibration Equipment: Before starting the calibration, verify that the calibration equipment is properly calibrated and within its specified accuracy limits. If necessary, perform any adjustments or recalibrations to ensure accuracy.

  3. Setup: Connect the pressure switch to the calibration equipment, ensuring proper sealing and secure connections to prevent leaks. Follow the manufacturer's instructions for the correct installation and setup of the pressure switch.

  4. Zero Adjustment: If applicable, perform a zero adjustment on the pressure switch to ensure that it reads zero pressure when no pressure is applied. Follow the manufacturer's guidelines for zero adjustment procedures.

  5. Pressure Application: Apply pressure to the pressure switch within the specified range using the calibration equipment. Gradually increase or decrease the pressure in small increments while monitoring the switch's response.

  6. Switch Point Calibration: Determine the switch points (setpoint, reset point) of the pressure switch by observing when it transitions between states (e.g., from open to closed or vice versa) as pressure changes. Record the pressure values corresponding to each switch point.

  7. Accuracy Verification: Verify the accuracy of the pressure switch by comparing its switch points to the expected values based on the applied pressure and the switch's specifications. Calculate the deviation from the expected values and determine if adjustments are necessary.

  8. Adjustment (if required): If the pressure switch's switch points deviate from the expected values beyond acceptable limits, make adjustments as needed to bring them into alignment. Follow the manufacturer's instructions for making adjustments, which may involve mechanical or electrical adjustments.

  9. Final Verification: After making adjustments, repeat the pressure application and verification steps to ensure that the pressure switch now meets the specified accuracy requirements. Make any additional adjustments if necessary.

  10. Documentation: Document the calibration process, including the equipment used, pressure values applied, switch points observed, any adjustments made, and the final calibration results. Keep records of the calibration for traceability and compliance with quality standards.

  11. Certification: If required, issue a calibration certificate indicating that the pressure switch has been calibrated and meets the specified accuracy requirements. Include relevant information such as calibration date, equipment used, and calibration results.

  12. Post-Calibration Testing: After calibration, perform additional testing or functional checks on the pressure switch to ensure that it operates correctly within its intended application.

By following these steps, you can effectively calibrate pressure switches to ensure their accuracy and reliability in detecting and controlling pressure levels according to the specified requirements.