How to calibrate displacer level transmitter?

1 Posts
1 Users
0 Reactions
662 Views
Posts: 18330
Admin
Topic starter
(@click2electro)
Member
Joined: 4 years ago

Calibrating a displacer level transmitter involves verifying its output signal response to changes in liquid level and adjusting it to match the desired measurement range. Here's a general procedure for calibrating a displacer level transmitter:

  1. Gather Equipment:

    • Obtain a reference measurement device such as a sight glass, dipstick, or another level transmitter with known accuracy.
    • Ensure you have a pressure source or control valve to vary the liquid level in the vessel.
    • Prepare a multimeter or loop calibrator to measure the output signal of the displacer level transmitter.
  2. Isolate the Transmitter:

    • Disconnect the displacer level transmitter from the process to avoid interference during calibration.
    • Ensure that the liquid level in the vessel is stable and not subject to sudden changes during calibration.
  3. Establish Calibration Points:

    • Identify the desired calibration points, including the minimum and maximum levels to be measured by the displacer level transmitter.
    • Use the reference measurement device to verify the actual liquid levels corresponding to these calibration points.
  4. Apply Test Pressure:

    • Adjust the liquid level in the vessel to the minimum calibration point.
    • Measure the output signal of the displacer level transmitter at this level using the multimeter or loop calibrator.
    • Record the measured output signal and the corresponding liquid level from the reference measurement device.
  5. Adjust Zero Point:

    • If necessary, adjust the zero point of the displacer level transmitter to match the measured output signal at the minimum calibration point.
    • Some displacer level transmitters have zero adjustment screws or potentiometers that can be used to fine-tune the zero point. Follow the manufacturer's instructions for adjustment.
  6. Apply Test Pressure for Span Calibration:

    • Adjust the liquid level in the vessel to the maximum calibration point.
    • Measure the output signal of the displacer level transmitter at this level using the multimeter or loop calibrator.
    • Record the measured output signal and the corresponding liquid level from the reference measurement device.
  7. Adjust Span:

    • If necessary, adjust the span of the displacer level transmitter to match the measured output signal at the maximum calibration point.
    • Some displacer level transmitters have span adjustment screws or potentiometers that can be used to adjust the span. Follow the manufacturer's instructions for adjustment.
  8. Verify Calibration:

    • After adjusting the zero and span, repeat the calibration process to verify that the displacer level transmitter accurately measures the liquid level within the desired range.
    • Ensure the output signal of the transmitter corresponds correctly to changes in liquid level.
  9. Finalize Calibration:

    • Once satisfied with the calibration results, finalize the process by documenting the calibration procedure, including the calibration points, measured output signals, adjustments made, and any observed deviations.
    • Label the displacer level transmitter with the calibration date and any relevant information.
    • Reconnect the transmitter to the process and ensure it operates correctly under normal conditions.
  10. Regular Maintenance:

    • Schedule regular inspections and calibrations of the displacer level transmitter to ensure continued accuracy and reliability.
    • Monitor the performance of the transmitter over time and recalibrate as necessary to maintain proper operation.

By following these steps, you can effectively calibrate a displacer level transmitter to ensure accurate and reliable level measurement in your process or application. If you're unsure about any steps or require assistance, consider consulting the transmitter manufacturer or a professional calibration service.