"Pressure switch calibration" is a critical process to ensure the accurate operation of pressure switches in various industrial applications. These switches play a crucial role in monitoring and controlling pressure-related processes. Let’s dive into the procedure for calibrating pressure switches:

Pressure Switch Calibration Procedure

Understanding Pressure Switch Basics

Before we proceed with calibration, let’s review some essential terms related to pressure switches:

  1. Set Point:
    • The pressure at which the switch changes state (opens or closes).
  2. Normal State:
    • The switch state when at barometric pressure or the ideal state (either OPEN or CLOSED, depending on the switch type).
  3. Reset Pressure:
    • The pressure at which the switch resets back to the normal state.
  4. Tolerance:
    • The allowable variation from the set point pressure.
  5. Repeatability:
    • The closeness of agreement between consecutive measurements.
  6. Dead Band:
    • The pressure difference between the change-of-states (e.g., OPEN to CLOSE or CLOSE to OPEN).
  7. Trip Type:
    • The direction in which the change-of-state occurs.
      • If the trip type is low, the change-of-state happens when the pressure is falling.
      • If the trip type is high, the change-of-state happens when the pressure is rising.

Calibration Steps:

  1. Depressurize and Isolate:
    • Ensure the pressure switch is depressurized and isolated from the process.
  2. Connect Pressure Source and Calibrator:
    • Connect the pressure source (preferably a hand pump) and the pressure calibrator (pressure gauge) to the pressure switch input.
  3. Check Electrical Output:
    • Use a multimeter in continuity mode to verify the switch contacts:
      • For a normally open (NO) switch, check that the contacts are open.
      • For a normally closed (NC) switch, check that the contacts are closed.
  4. Test Set Point:
    • Apply pressure to the switch until it changes state (e.g., from OPEN to CLOSE for an NO switch).
    • Record the pressure at which the switch changes state (set point).
  5. Test Reset Pressure:
    • Release pressure until the switch resets back to the normal state (e.g., from CLOSE to OPEN for an NO switch).
    • Record the reset pressure.
  6. Repeat for Repeatability:
    • Repeat the test at least once to determine repeatability.