What is zero stability on a Coriolis flow meter? How to do zero calibration check of Coriolis flow meters?
Zero stability in a Coriolis flow meter refers to the ability of the meter to maintain a stable output signal (zero flow rate) when there is no flow through the meter. It indicates the degree of drift or fluctuation in the zero flow rate reading over time. Zero stability is an important parameter as it ensures accurate measurement of flow rates, especially at low flow rates or during periods of no flow.
To perform a zero calibration check of a Coriolis flow meter, follow these general steps:
  1. Ensure Zero Flow Conditions: Before starting the calibration check, ensure that there is no flow through the Coriolis flow meter. This can be achieved by closing any valves upstream and downstream of the meter and confirming that the flow rate reading is zero on the meter's display or transmitter.
  2. Access Calibration Function: Most Coriolis flow meters have a built-in calibration function accessible through the meter's display or configuration software. Refer to the manufacturer's manual or documentation for instructions on how to access the calibration menu.
  3. Initiate Zero Calibration: Once in the calibration menu, locate the option for zero calibration or zero adjustment. Select this option to initiate the zero calibration process.
  4. Wait for Stabilization: Allow sufficient time for the flow meter to stabilize and settle at zero flow conditions. This typically involves waiting for a specified period (e.g., several minutes) to ensure that any transient effects or fluctuations in the zero flow rate reading have dissipated.
  5. Confirm Zero Calibration: After stabilization, verify that the zero flow rate reading on the meter's display or transmitter remains stable and consistent. If the reading fluctuates or drifts significantly, it may indicate a need for zero calibration adjustment or troubleshooting.
  6. Perform Zero Calibration Adjustment (if necessary): If the zero flow rate reading deviates from the expected value, you may need to perform a zero calibration adjustment using the calibration function. Follow the on-screen prompts or manual instructions to adjust the zero offset and align the meter's output signal with the true zero flow rate.
  7. Document Calibration Results: Once the zero calibration check is complete and any necessary adjustments have been made, document the calibration results, including the date, time, and any corrective actions taken. This documentation is essential for quality assurance and compliance with regulatory requirements.
  8. Repeat Periodically: Zero calibration checks should be performed periodically as part of routine maintenance and calibration procedures to ensure the ongoing accuracy and reliability of the Coriolis flow meter. The frequency of calibration checks may vary depending on factors such as the stability of the process, environmental conditions, and manufacturer recommendations.
It's important to note that the specific steps and procedures for zero calibration checks may vary depending on the model and manufacturer of the Coriolis flow meter. Always refer to the manufacturer's documentation and guidelines for accurate and reliable calibration procedures. Additionally, consider consulting with qualified personnel or calibration specialists if you are unsure about performing calibration checks or adjustments.